Test Equipment: Oscillocopes
Started by MarkE, February 14, 2015, 04:35:20 PM
Brian516
yeah, the trigger holdoff is causing me some weirdness.
Uploading a vid of it now...
https://www.youtube.com/watch?v=-QnIyMVbbHE
Those type of pots that are right up against the faceplate I haven't cleaned at all.  On all the other ones, of which were mostly noisy, I dripped a little Deoxit D5 on the shaft after cleaning the gunk/fuzz off, and worked them back and forth until the stuff got inside, then put another drop and repeated.  did this with 2 or 3 drops on each pot and all of those seem to be working fine.   Maybe I should start by doing that on the rest of these pots? Or do you think that it's most likely something else?
Oh, and the cover is off.
Brian516
Ok, I measured the voltage at the main power supply test points, and all the other ones.  Had some interesting results on the Vertical preamp board....
(TP1518   +110V)       110.3VDC
(TP1536   +55V)          54.93VDC
(TP1548     +15V)         14.95VDC         1.34VAC
(TP1558     +5V)          4.999VDC
(TP1568     -8V)           7.97VDC
TP1486    17.77V
TP1590    10.74V    3.264VAC
TP1594     9.54V     2.008VAC
Vertical Preamp Board:
TP162      started at .6mV then climbed to 2.6mV then dropped back down to .2mV and kept going back and forth.
          158.3mVAC
TP164      back and forth between 13.6mV and 14.2mV
          162.9mVAC
TP346  back and forth between 23.6mV and 29.2mV
          113.6mVAC
TP322     -29.7mV
          back and forth between 1.463mVAC and 1.490mVAC
TP374     -1.175V
TP324     -66.4mV
          46.2mVAC
TP364     -1.174V
TP262     back and forth between -18.8V and 19.8V
          75.8mVAC
TP147     8.35V
TP141     8.37V
TP247     8.33V
TP241     8.33V
Is it possible that the AC riding on the +15V rail is causing all of this instability and AC on all these test points?   seems like there is a good bit of AC running rampant in my scope.    I'm gonna start inspecting the schematics to see if I can narrow down what component it could be.   I'm sure you guys will be able to tell me exactly where to look, though.
Brian
Hmm, interesting.   I adjusted the 55V adjust pot to EXACTLY 55V, and now there is absolutely no AC on the +15V rail.    or any of the others.
I am going to go thru and check everything else for stability and AC now....
Some new numbers after the +55V rail adjust and disappearance of the AC on the +15V rail:
(A9 board)
TP1486    18.38V (the manual says that this is linked to the Intensity adjustment and varies from 0V to 75V)
TP1590    9.54V
          3.268VAC
TP1594    10.76V
          2.011VAC
TP1423 (-2450V rail)   No AC.   My meters max out at 1000V so I'm not checking the DC at that point.  Wish I could.... maybe a X10, or even a X3 probe can be made for a DMM??  I don't see why not, but in order for it to be accurate I would have to use exact value components, and add in a tuning cap and maybe small VR? hmm.....  could come in handy if it can be made......
(Vertical Preamp board)
TP162     3.3mV to 4.5mV
          158.7mVAC
TP164     12mV to 8.8mV
          163.3mVAC
TP346     -32.6mV to -25.6mV
          114.7mVAC to 114.3mVAC (rapidly changing)
TP322     -32.2mV to -31.3mV
          4.60mVAC
TP374     -1.178V
          2.7mVAC to 4.4mVAC
TP324     -69.3mV
          46.3mVAC
TP364     -1.177V
          0mVAC to 1.4mVAC
TP262     -23.5mV to -24.5mV
          76mV
TP147     8.37V
TP141     8.38V
TP247     8.35V
TP241     8.34V
The manual says that these are what the values of these test points should be
TP147     8.5V
TP162     0V
TP141     8.5V
TP241     8.4V
TP247     8.4V
TP262     0V
These two test points are in the Calibrator circuit.  The values I am getting are way under the values that the manual shows I'm supposed to have...
(Page 234 of the Tek 465 manual linked to by TinselKoala on page 5 or 6?)
TP1590     10.8V
TP1594     54.7V
This circuit draws it's power from the 55V, 15V, and 5V rails, so if my values are off it must be one of the components in that circuit that is damaged, correct?
I am going to hook up to the closest ground and check to see if the values match the specified ones in exactly the places where the arrows are pointing to.
All of the rails appear to be doing great, especially now that I have gotten rid of the AC on the 15V.  I have had it running for the last several hours non stop except for a couple of minutes and everything is good.  Before I adjusted the 55V adjuster, I had adjusted the "A trigger Level" and "A trigger holdoff" to try and keep the trace steady constantly, but it still glitched a little from time to time. However after adjusting the 55V, it hasn't glitched a single time that I have noticed. :)
There are still definitely some issues to be worked out, though.   I think here after I check a few spots for their specified voltages, I am going to pull the knobs on the A and B trigger level/slope and the A trigger holdoff and put a couple of drops of Deoxit on the shaft and let it sit, then work it around in the morning and maybe put one more drop in, too, just to "rinse" off any crud.  I can't get the stuff out of there, so I can't be putting too much in, obviously.
I shall attempt to isolate all the other issues according to the manual as soon as I get a chance to tomorrow.    I'm going to start finding deals on all the stuff listed in the "calibration" section that I will need for a performance check first of all, and then calibration.  I will have to wait until I can pick up a good quality function gen that has all the functions I need in the tolerances I will need them in. I'm sure that won't be a problem.   Oh, and a good Variac....
Brian
Ah, poop.... I'm going to have to redo all of that again tomorrow WITHOUT anything hooked to the input..................  ::)
picowatt
Brian,
Regarding the trigger holdoff, for repetitive signals such as the square wave from the calibrator, you will use the minimum holdoff setting of that control.  Set it there and forget about it for now.
The trigger level control should be able to be set for stable triggering in both positive and negative going slope selection, AC, DC, HF reject, etc when looking at the calibrator.
Regarding the ripple disappearing when the 55V rail was adjusted.  As per the manual, that rail was already within spec.  Possibly the Vadjust control was noisey, but it is difficult to grasp how adjusting it cleaned up the ripple issue.  When you get a probe, you can use the scope to look at the supplies for ripple.
Regarding the calibrator TP measurements,  what is the level that you are seeing on the screen when connected to the input channels?  I think it is supposed to be around 300 milivolts PP and if it is, move on.  Keep in mind that the calibrator is typically used as only a fairly rough reference regarding amplitude and frequency.  Its main function is to provide a fairly fast risetime signal to calibrate (adjust) the probe's compensation cap (when you get a probe).
Other than verifying the main supplies, as you have done, I would suggest using the scope or going thru the performance verification part of the manual to see what all is and is not working.  A full calibration is going to require a few pieces of equipment you may not have.  As well, it must be done in the order given in the manual, which typically begins with power supply rails.
You can do some DC measurements to verify the channel inputs are close.  Use a 9V battery, connect it to an input channel, see what it reads using the scope and verify that voltage with your DVM.  Repeat for other channel.  This will give you at least an idea of where the preamp inputs are.  You can repeat using a lower V battery to check accuracy at various voltages.  Of course, if you have an adjustable DC supply, use that instead.
Don't start randomly tweaking cal trims in the scope unless you are setup to do the cal correctly with the required equipment.  You may be able to do DC offset adjust and some other cals but I would use the scope a bit and see what is or is not working well.
Do you have a signal/function generator?
PW
@Brian: To display both traces with signals from the calibrator, you don't need to use different V/div settings, you simply use the channel's Vertical Position control to put one trace on the upper part of the screen and the other trace on the bottom part of the screen. Set Vert Mode to "Alt" . Set the CH1 input coupling to "ground" to set the baseline on some graticule line in the upper part of the screen, then turn the channel input coupling back to "DC" or "AC" and scale the trace using the V/div control so it just spans a couple of vertical divisions. Do the same with CH2 except use some graticule line on the lower part of the screen for the "ground" or zero volts baseline. Then you'll have two traces that don't overlap (unless you scale them to do so using the V/div control) and since both will have the same setting on the V/div control you can compare their amplitudes, which should be the same . (Red "var" knob on the V/div controls fully CW and in the detent.)
Quote from: Brian516 on February 22, 2015, 06:54:01 PM
OK< I see that you have figured out how to show both traces properly separated using the vert position controls. It doesn't look "weird" to me now. I think the issue you are seeing now has to do with the trigger _level_ setting rather than the trigger holdoff. For this signal turn trigger holdoff to the "norm" position and leave it there.
Make sure Trigger Source is on CH1 or CH2, mode is AUTO and adjust A Trigger Level for a stable display. Horiz Display on "A".
Try turning the horizontal timebase to a faster setting so you are only displaying three or four cycles of the waveform on the screen. This will help with some diagnostics, now that you are getting traces there is no need to keep the scope in the 1 ms/div setting.
Anyhow, it does look much better now, perhaps your cleaning did indeed fix it! Or maybe it just likes running on its side.   ;)
Running a full calibration is probably overkill.  It would be nice if you could use some different input signals, like from your GenRad oscillator, for some higher-frequency testing.
It is likely that the little ripple you are seeing is due to the casual grounding we have been using for the input signals, or perhaps some ripple on a power supply due to a faulty capacitor.